Will the real slim shady distortion plot please stand up?
Can someone explain to me the reason for the differences? My cursory glance (non- mathematical/statistical trained brain) says it's related to noise polluting the measurement. But why?
The measurement noise floor is determined by the energy in the stimulus versus the background noise level. Longer sweeps and more repetitions increase the sweep energy, gaining about 3 dB for each doubling in either length or reps, but the energy is still spread out over the span of the sweep. Stepped sine puts all its energy into each stimulus step and excludes all noise outside the bins of the fundamental and harmonics, so gets very good signal to noise. It can be more prone to reflections though, probably worth putting the mic a lot closer to minimise their contribution.
You can overlay the results for individual harmonics on the Overlays Distortion graph, by the way.
Windows 11.
Sorry, I’m not certain I understand “checkboxes look huge”
Is it the Y axis of 20dB per major division?
I hear what you’re say about putting the mic closer. In practice I have to balance being able to get the mic close enough to minimise reflections, but not too close such that I overload the mic. As you know, if we’re trying to measure the distortion of low distortion drivers, we need to be careful that the microphone is used in a range where it’s own inherent distortion is lower than the Device Under Test.
The microphone in the test above is the Sonarworks Xref20 v4 (2015 version) whose maximum SPL is quoted at 128dB (? % THD) This puts the microphone’s ideal range to well under 100dB (exactly how much- TBC) if I’m trying to characterise an undocumented driver with -60dB H2 and -70dB H3.
I will need to substitute another mic before I move it closer eg. 20cm, and test at 2.83V.
For now, do you agree that the bottom graph is the most accurate representation?
Have a peaceful Easter!
PS.
Thank you for tip on the Overlays Distortion graph. I wasn’t even aware of this feature. Another donation coming...
The stepped sine should be most accurate, but the dips look suspicious to me hence the comment about reflections. Might be worth doing a run with a higher PPO to see how localised those dips are.
That workstation is a 4K screen running at 125% zoom and screen capture 800 pixels horizontal.
When it's rendered on a 1080 screen captured at 1024 pixels it looks like this. Not sure why, But let's overlook that for now.
Here's the 24ppo stepped sine:
Moving forward I will use log sine sweep for to look at frequency response and stepped sine for distortion.
Perhaps not 512K FFT though...it's the audio equivalent of watching paint dry.
Cal files are used by the RTA, they are optional for sweep measurements:
If the Analysis PreferenceApply cal files to distortion is selected the results will include corrections for the cal file responses (as is the case for the RTA distortion figures). Applying the cal files provides more accurate results in regions where the fundamental or harmonics are affected by interface roll-offs but boosts the noise floor in those regions. This should be borne in mind when viewing the results. If large cal file corrections are required make sure the Analysis PreferenceLimit cal data boost to 20 dB is not selected. Note that any subsequent changes to the cal files will NOT update the distortion results, they are generated from the cal files that were in use at the time the measurement was made.
It turns out, most of the distortion I am measuring is in fact the microphone distortion....
Let me explain why:
The microphone in the test above is the Sonarworks Xref20 v4 (2015 version) whose maximum SPL is quoted by the manufacturer at 128dB. Although the manufacturer doesn't list how this max SPL is determined, it is possible that it is a 3% THD rating (TBC). To measure distortion correctly, we want to measure the speaker/transducer under test, NOT the microphone's own self-distortion.
Based on past findings, it appears that electret condenser microphones have a 2nd order distortion behavior.
eg. 3% THD (-30dB) -> 1% THD (-40dB) -> -10dB
1% THD to 0.3% THD -> -10dB
0.3% THD to 0.1% THD -> -10dB
X-axis is SPL in dB; Y-axis is harmonics in "%", dots are experimentally measured values of THD in Panasonic WM-61.
Reference:
В теме дается обзор схемотехнических решений, повышающих перегрузочную способность электретных микрофонов по допустимому звуковому давлению (SPL), и предлагается для опробования простая, но эффективная и неприхотливая схема предусилителя, снижающая (в симулировании) нелинейные искажения его...
forum.vegalab.ru
Based on this, the WM61 has ~0.05% (-66dB) at 115dB
Although extrapolation of the purple line suggests it may be able to measure at lower distortion, AFAIK, this has not been verified (grey dots). Furthermore, the WM61 has a self-noise of 32dB.. which limits it's abiliity to resolve harmonics much below this noise floor.
Another one:
Reference:
Production Partner.de's review of iSEMcon EMX-7150 microphone (06/2012 Issue)
This microphone looks like it can resolve H2 (k2 in chart) as low as -70dB, when it is observing 115dB.
As we can see, below 115dB, H2 actually goes up to -65dB. This suggests it's noise limited. Indeed the specification for the microphone is "~30dB(A)"
So for this mic the sweet spot of measuring H2 this mic is around 100-105dB, but for H3, perhaps 100-120dB/
Finally there is the B&K microphones:
Reference:
Brüel & Kjær Microphone Handbook Vol 1: Theory (2-42)
Here's the 4191:
Let us extend this the 2nd harmonic distortion line to the left: We see that the 0.01% (-80dB) distortion point is 116dB for the 2nd harmonic.
116-80 = 36dB.
Well this might have a chance because the self noise of the mic/pre-amp combination is lower, at 21.4dB(A)
So if a true condenser microphone has a maximum SPL rating of x dB @3% distortion, it's self distortion of 0.01% may be found at 50dB lower than it's maximum SPL rating
As there was a lot of discussion about microphone distortion here recently and I'm actually working on a pressure chamber which can produce >160dBSpl ... I spent an afternoon to do some microphone measurements. Setup A pressure chamber is a device with a speaker membrane with a very small and...
www.diyaudio.com
This puts the Sonarworks XRef20 (2015-2023 models) optimal operating range for distortion measurements to MUCH LOWER than the maximum SPL rating.
As you can see from my measurements, the microphone was observing to levels of up to 102dB when measuring the tweeter at 31.6cm. So the distortion measurement is NOT just of the driver, but includes that of the microphone. How much is the tweteer and how much is the microphone has not yet been determined. (TBC)
And finally, a quick and dirty way to check to see if you microphone is contributing to the distortion is if the H2 has the same shape as the fundamental.
Although the Stepped Sine is more accurate, the log sine sweep is not far off. Below 1KHz they are very similar. Above 2KHz the Stepped Sine can resolve the noise floor and thus higher order harmonics better than the Sweep. However it is very time consuming.
Are there settings within the Stepped Sine function that can bring this test to around 2-3 mins?
Assuming I still want to test 20Hz to 20KHz, 24 ppo. And still allow one to resolve higher order harmonics down to below -80dB? (ie. At least equal or better than Sweep)
The test time is affected by the buffer lengths, since REW must wait for a frequency change to pass through the replay and record buffers. If you try ASIO the buffers are much smaller (even at their largest settings) than for Java drivers and so the test is quicker.