Just circling back to on this.
It turns out, most of the distortion I am measuring is in fact the
microphone distortion....
Let me explain why:
The microphone in the test above is the Sonarworks Xref20 v4 (2015 version) whose maximum SPL is quoted by the manufacturer at 128dB. Although the manufacturer doesn't list how this max SPL is determined, it is possible that it is a 3% THD rating (TBC). To measure distortion correctly, we want to measure the speaker/transducer under test, NOT the microphone's own self-distortion.
Based on past findings, it appears that electret condenser microphones have a 2nd order distortion behavior.
eg. 3% THD (-30dB) -> 1% THD (-40dB) -> -10dB
1% THD to 0.3% THD -> -10dB
0.3% THD to 0.1% THD -> -10dB
X-axis is SPL in dB; Y-axis is harmonics in "%", dots are experimentally measured values of THD in Panasonic WM-61.
Reference:
В теме дается обзор схемотехнических решений, повышающих перегрузочную способность электретных микрофонов по допустимому звуковому давлению (SPL), и предлагается для опробования простая, но эффективная и неприхотливая схема предусилителя, снижающая (в симулировании) нелинейные искажения его...
forum.vegalab.ru
Based on this, the WM61 has ~0.05% (-66dB) at 115dB
Although extrapolation of the purple line suggests it may be able to measure at lower distortion, AFAIK, this has not been verified (grey dots). Furthermore, the WM61 has a
self-noise of 32dB.. which limits it's abiliity to resolve harmonics much below this noise floor.
Another one:
Reference:
Production Partner.de's review of iSEMcon EMX-7150 microphone (06/2012 Issue)
This microphone looks like it can resolve H2 (k2 in chart) as low as -70dB, when it is observing 115dB.
As we can see, below 115dB, H2 actually goes up to -65dB. This suggests it's noise limited. Indeed the specification for the microphone is "~30dB(A)"
So for this mic the sweet spot of measuring H2 this mic is around 100-105dB, but for H3, perhaps 100-120dB/
Finally there is the B&K microphones:
Reference:
Brüel & Kjær Microphone Handbook Vol 1: Theory (2-42)
Here's the 4191:
Let us extend this the 2nd harmonic distortion line to the left: We see that the 0.01% (-80dB) distortion point is 116dB for the 2nd harmonic.
116-80 = 36dB.
Well this might have a chance because the self noise of the mic/pre-amp combination is lower, at 21.4dB(A)
So if a true condenser microphone has a maximum SPL rating of x dB @3% distortion, it's self distortion of 0.01% may be found at 50dB lower than it's maximum SPL rating
Reference:
As there was a lot of discussion about microphone distortion here recently and I'm actually working on a pressure chamber which can produce >160dBSpl ... I spent an afternoon to do some microphone measurements. Setup A pressure chamber is a device with a speaker membrane with a very small and...
www.diyaudio.com
This puts the Sonarworks XRef20 (2015-2023 models) optimal operating range for distortion measurements to
MUCH LOWER than the maximum SPL rating
.
As you can see from my measurements, the microphone was observing to levels of up to 102dB when measuring the tweeter at 31.6cm. So the distortion measurement is NOT just of the driver, but includes that of the microphone. How much is the tweteer and how much is the microphone has not yet been determined. (TBC)
And finally, a quick and dirty way to check to see if you microphone is contributing to the distortion is if the H2 has the same shape as the fundamental.