Hi John,
I've been comparing REW 5.40 beta 135 with QA40x 1.224, using the same QA403 to measure an amplifier at approximately 1 W into 8 Ω.
I performed two separate comparisons:
1. SNR measurements (generator ON)
With a nominal 1 kHz test signal at 192 kHz, QA40x reported an SNR of 99.08 dB, while REW reported 94.4 dB, a difference of approximately 4.7 dB. The measured amplifier gain was nearly identical: 27.45 dB and 27.46 dB, respectively.
2. Noise measurements (generator OFF)
With the generator disabled but still physically connected, QA40x measured −89.14 dBV and REW measured −90.17 dBV, a difference of only 1.03 dB.
These are separate measurements: the generator-OFF readings represent noise levels in dBV, whereas the SNR figures express the ratio between signal and noise.
Importantly, the original SNR comparison was made before we discovered the ASIO sample rate startup issue. REW was actually generating 250 Hz instead of 1 kHz, so those SNR results need to be repeated at the correct frequency.
Once the sample rate is verified, I'd like to understand whether REW and QA40x use different SNR calculation methods or measurement bandwidths that could explain any remaining difference.
Thanks!
I've been comparing REW 5.40 beta 135 with QA40x 1.224, using the same QA403 to measure an amplifier at approximately 1 W into 8 Ω.
I performed two separate comparisons:
1. SNR measurements (generator ON)
With a nominal 1 kHz test signal at 192 kHz, QA40x reported an SNR of 99.08 dB, while REW reported 94.4 dB, a difference of approximately 4.7 dB. The measured amplifier gain was nearly identical: 27.45 dB and 27.46 dB, respectively.
2. Noise measurements (generator OFF)
With the generator disabled but still physically connected, QA40x measured −89.14 dBV and REW measured −90.17 dBV, a difference of only 1.03 dB.
These are separate measurements: the generator-OFF readings represent noise levels in dBV, whereas the SNR figures express the ratio between signal and noise.
Importantly, the original SNR comparison was made before we discovered the ASIO sample rate startup issue. REW was actually generating 250 Hz instead of 1 kHz, so those SNR results need to be repeated at the correct frequency.
Once the sample rate is verified, I'd like to understand whether REW and QA40x use different SNR calculation methods or measurement bandwidths that could explain any remaining difference.
Thanks!