You can do a substitution calibration, making a measurement with each mic in turn from exactly the same position (usually best close to a source) and then using the All SPL trace arithmetic A/B to calculate the ratio of the responses. Probably best using magnitude-only data for the ratio, which you can make from the measurements by selecting each in turn (with no other measurement selected) and using the "Average the responses" button which will produce a magnitude-only copy. The UMIK-1 measurement needs to use the UMIK-1 cal file. Levels should be manually aligned to match at 1 kHz.
The regularisation option for division would help avoid issues due to dips in the reference response, but they are best minimised through the measurement setup and windowing.Since v5.20.13 there are new trace arithmetic features which may be even better for microphone calibration? Yes/no/maybe?
RMS average is probably best.I assume that it is a good idea to average multiple measurements instead of just using one take, but which averaging method is most suitable? There are "dB" / "Vector" / "RMS" Average buttons available? (some searching gave me paper "The Geometric Mean is a Superior Frequency Response Averaging Method for Human Body Vibration" on ResearchGate, which unfortunately did not make me any smarter...)
Trace arithmetic is done using the windowed responses, so just make sure the windows are as you wish them to be set before doing the arithmetic. You can see the effect of window changes interactively on the Impulse graph by dragging the window indicators at the top of the graph, only the right hand window is relevant. There is a button on the IR windows dialog to apply window settings to all measurements.Windowing should be used to filter out reflections? Any tips to make sure that windowing is included in trace arithmetic?
REW has plenty of smoothing options.Finally a new calibration file probably needs some smoothing applied, as well as it needs some reasonable amount of data points in the text-file. Is that handled by REW somehow or should be done in other software? I can only think of writing a Python script, but that again involves many arbitrary decisions which may or may not be optimal (trying to "do it right"...).
There isn't a lot to explain, it is just like ordinary arithmetic except it is carried out point by point on the complex frequency response.In addition I would appreciate some links to extra reading or more detailed explanation of trace arithmetics. I failed to find some resource that explains the notation and practical meaning behind "A/B", "1/A", "A-B", etc
Google is your frenemy... https://www.roomeqwizard.com/help/help_en-GB/html/graph_allspl.html... Description of Trace Arithmetic and caveats of usage are a few sections down...5. In addition I would appreciate some links to extra reading or more detailed explanation of trace arithmetics. I failed to find some resource that explains the notation and practical meaning behind "A/B", "1/A", "A-B", etc.
The regularisation option for division would help avoid issues ...

RMS average is probably best.
So... expanding on the John's response, Here is my draft procedure for microphone calibration:
Objective: to verify and improve calibration of one measurement microphone against another, reference microphone.
The concept: we find the OFFSET correction from the measurement with OLD calibration file to the REFerence microphone measurement. Then we apply the offset to the old calibration file to get the NEW calibration file. (this way low-bass calibration and gain can be re-used from the original file in case it was not possible to reliably re-calibrate those parameters in your setup).
Steps:
- Record speaker measurements with both REFerence microphones and microphone that needs to be re-calibrated - "DUT" - device under test..
- Use substitution of microphones in front of the same speaker. Every microphone is placed as-precisely-as-possible (precision within a couple of millimeters) in the same spot, one at a time. (A lot of people ignore this and place microphones even several centimeters off from one another which is a Bad idea)
- Speaker should be physically small, coaxial or a "full-range" driver. It does not need to be flat, just needs to have enough output at all relevant frequencies to give enough dynamic range versus ambient noise and microphone self-noise. Alternatively, the calibration will only be valid at frequencies that speaker can reproduce (it is rare to get a speaker with good lower bass output that is otherwise compact and full-range).
- Use typical Measurement Sweep at 48 or 44.1 kHz (perhaps using longer sweep length 512K or if possible, a few repetitions).
- Check the maximum THD distortion in the measurement. Consider to not use frequencies with > 3-5% THD. (If distortion is high - either the speaker is pushed too loud or the speaker can not produce enough output to overcome ambient noise - typical at low frequencies).
- + other optimizations to reduce impact of reflected sound, avoid comb-filtering from multiple drivers playing the same frequencies, leave enough distance for sound of multiple drivers to integrate, as well as keep direction of direct sound as close to point-source as possible.
- (optional?) apply "IR Windows" to all measurements. (Recommendations not clear at the moment, but to-start-with try frequency dependent window with 15 cycles, Tukey 0.25)
- Apply light smoothing on the data, for example 1/48. (The smoothing and windowing will be used in the Trace arithmetic.)
- Alternatively, the smoothing may be redundant if you already applied "IR Windows".
- Apply OLD calibration file to the DUT measurement data. (Note that it is possible to add or replace microphone calibrations for each measurement at any point. Also, you may have applied the calibration already when you selected microphone for measurement in REW Preferences).
- Run "Average The Responses" (the same as "RMS Average") on the DUT data. (To both bake-in the old calibration file and remove phase information in the DUT data).
- Do the same procedure for the REFerence data: apply any calibration files and then take "RMS Average" of the REFerence microphone data. (To both bake-in the calibration file and remove phase information. Note, if you can, take an average of multiple reference microphones - no microphone is perfect, so average of several top-quality microphones is better than any single one of them.)
- Use "Align SPL..." on the averages of DUT and REF.
- If you are calibrating miniDSP UMIK then "Align SPL" to 0db at 1 kHz with 0 or 1 octave span (because UMIK calibration files are always 0dB at 1000 Hz).
- For other cases, choose alignment parameters that make the most sense (usually using midrange frequencies within 250 to 2000 Hz, because at high frequencies reflections make the data less consistent between microphones).
- Use "Trace arithmetic" ==> "A/B" where you divide: averaged DUT data / REFerence data. As a result you get the "OFFSET correction".
- Load in the OLD calibration file as a measurement (using File => Import frequency response).
- Apply 1/3 (or 1/6) smoothing on the "OFFSET correction" to make it about as detailed as the OLD calibration curve. (The smoothing will be used in the Trace arithmetic.)
- Use "Trace arithmetic" ==> "A*B" where you multiply: smoothed OFFSET correction * OLD calibration file. As a result you get the initial "NEW calibration". (Note, the input data has no phase and there is no need to do SPL alignment before multiplication).
- Manually align SPL of the NEW calibration by going to "Measurement actions" and dialing in "SPL offset" value to match OLD calibration and press "Add Offset to data". Alignment is done at a frequency where you plan to join the OLD and NEW calibration data together (a reasonable value could be 1 kHz, depending on the speaker size, room reflections and differences between old and new calibration curves).
- Note: to make the manual alignment easier: first use "Align SPL..." function to 0 dB. and then do the manual "SPL offset" in fully zoomed in view.
- Export NEW calibration as text with 48PPO and NO-additional smoothing (smoothing should already be sufficient from previous operations).
- Check that you get the values all the way to 20 kHz. If not, simply export one more text file with "Use measurement resolution" setting and copy the last row out of this "high-resolution" file into your normal 48PPO text file. (You can also round the frequency to "20000.0")
- Splice the data: copy the relevant frequency-SPL rows from the exported NEW calibration text file into the original OLD calibration text file using basic text editor. This way only valid values are replaced (for example low-bass calibration and gain can stay from the original file).
- Done.
I hope we can have a discussion and peer-review on this draft. I am trying to fix mistakes in this tutorial for myself and others.
In my testing, UMIK2 had more than 2dB error at high frequencies compared to Earthworks M23R reference (and I am quite confident in accuracy of my measurement setup and of the M23R). So this sort of calibration is quite significant in practice for people who use affordable USB microphones.
Hello, as there was not much response to expand on the tutorial in this forum, I posted a full, updated tutorial on a blog:@Sergej
Sergej, that is fantastic - any chance that you could do a video tutorial for the rest of us?